I just wanted to pick everyone's brains about what their general starting conditions are to see a wide variety of analytes when you don't have little knowledge of their functional groups or molecular weights.
I was looking for a trace impurity for which we had no information, nothing about MW or functional groups. I was able to see a small peak in UV at 258 nm but the total ion chromatogram showed nothing above the background. Since I was at least told that it was a small molecule, I started with a mobile phase of 10 mM ammonium acetate with a 5-95% methanol gradient through an 2.1 x 3 mm C18 column with a flow of 0.2 mL/min in APCI mode, looking at pos and neg with a scan from 100 - 1000 and fragmentor set to 70V. In my limited experience (2 years), it seems that you have to have to inject a concentrated sample to even see a peak in the total ion chromatogram. My boss (who isn't a chemist) thinks that mass spec is a cure-all and if you shoot anything in there you should see something.
So anyway, if anyone would like to share what their idea of a good general starting point is I'd appreciate it.