PROBLEMS WITH FID

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Everyone,
I have one 7890A GC, working, but in last week I found a problem :
first we don't have conection with software and machine because when I injected sample the time delay a few minutes to start.
second when the run is processing the output decrease at negative pA(11.1pA to -2.1pA) it seem like the detector is switch off, but if you abort it came to the intial baseline.Can help me please?

Just a few thoughts:

-Did you try to reboot the system (switch on and off both computer + GC) to re-establish the connection between the instrument and PC?

-Regarding the FID: It can be that your detector is dirty! Is the flame on (do you see condensation when placing a glass object above it) when in stand-by or when analyzing? Is the signal constant (no considerable fluctuations) when the instrument is in stand-by? I woud check if there is flow passing in the column (put detector end part in methanol and you should see bubbles) and then clean the detector.

Regards,
At_om
2 posts Page 1 of 1

Who is online

In total there are 93 users online :: 1 registered, 0 hidden and 92 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot] and 92 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry