We are having a very peculiar problem and request your help & suggestions to find an amicable solution to this problem.
Customer Sequence of sample is: Blank followed by a sequence of six standard injections
Problem: Always his first injection Area from the same Vial (6-replicate injections from the same Vial) is giving less or variable response leading %RSD of 4.6 (if we remove the first injection and consider the remaining injections the %RSD is around 1%) We also tried to remove the Septa assuming this to be the case of some Air getting sucked from the vial during the first injection but even this didn't workout.
Also repeated this set of sequence samples more than five times and each time the results from the first injection are the cause of %RSD failure
Please let us know what else needs to checked to find a solution.
Does FDA allow the enduser to deselect the first injection and consider the remaining 5-6 runs only for the %RSD calculations ??? if yes is there any documented statement which can be provided to circumvent the current situation.
Await for your further reply.
warm regards,
gsck9