We note the following problems:
- Connecting all the modules on the control panel is painfully slow.
- Not surprisingly, any command changes (i.e. changing flow rate, etc) take painfully slow to execute. This is frustrating when mobile phase is a precious resource.
- After a run, the software does not initiate injection of the next sample. It continues the same old run after UV acquisition is turned off and the sample run time has clearly ended. There is nothing wrong with the program. We are using the same program file as the other LC's and that is running without a glitch.
- Only the first 5 minutes of the chromatogram is recorded on the report, even when I see peaks generated on the acquisition display.
We suspect that there is a communication issue between the jetdirect LAN card and the chromeleon server. BootIP cannot find the IP address for the LAN. We set the IP address manually. Then tested it by pinging it via the C prompt. No response.
Now, I cannot even connect my modules- Chromeleon is not communicating via LAN.
Anybody have communication issues similar to ours? Any advice is appreciated! Thanks!!!!!