ISO9377-2 oil index or oil in water: septum bleed
Discussions about GC and other "gas phase" separation techniques.
8 posts
Page 1 of 1
Recently, we started to use ISO 9377-2 method for oil index in water (waste water) determination. System: Varian GC FID, 15mx0.25x0.25 VF1 column, on column injector. 5m of 0.53 precolumn. The problem is very high septum bleed - signal peaks in range of C10-C40 are in order of 30 mV (0.2 mg/l) and septum peaks are in the same order (going out ~150C. Temp. program: 50C for 2 min. and 60C/min to 320C hold for 2 min. Flow - 4 to 2 ml/min. I've tried changing almost everything, but nothing helps. And another big problem - C40/C20 ratio is very small = 0.35. It should be >0.8 according to standard. Maybe someone had the same problems? I will be glad to get any help.
-
- Posts: 18
- Joined: Sun Sep 12, 2004 5:33 pm
Dear Remirama,
I work with a method based in 9377-2, but I work in splitless mode, at 340°C, column flow 7.4 ml/min, dtector at 325 °C. My initial temperature is 40°C for 1.50 min, then up to 60°C at 5°C/min and finally up to 315°C at 15°C/min. In such conditions C40/C20 ratio is nearly or higher than 0.80 and no bleeding or not significant.
I hope this will help you
Carmen
I work with a method based in 9377-2, but I work in splitless mode, at 340°C, column flow 7.4 ml/min, dtector at 325 °C. My initial temperature is 40°C for 1.50 min, then up to 60°C at 5°C/min and finally up to 315°C at 15°C/min. In such conditions C40/C20 ratio is nearly or higher than 0.80 and no bleeding or not significant.
I hope this will help you
Carmen
-
- Posts: 3
- Joined: Wed Jan 26, 2005 11:37 pm
Thank you for your answer, but just to be sure two more questions. Are you turning on split after some time (~5 min)? Is injector isothermal (all the time 340 or temperature program)?
Now i'm working with such configuration: injector 40 to 340C, split is on initialy (vents out hexane, very nice peak), after 10s split is off for 2 min. and after on for remaining time. Injection volume - 2ul. Temp. program: 30C for 2 min, after to 330C, at 30C/min and holds for 2 min. C40 comes at 13-th minute (15mx0.25x0.25 column). C40/C20 = 0.87. With longer period of initial split on, discrimination appears for the low boiling point hydrocarbons.
I think, it is not possible to use on column injector for that analysis, at least for injectors that use hot oncolumn and do not have septum purge.
Best regards,
remi
Now i'm working with such configuration: injector 40 to 340C, split is on initialy (vents out hexane, very nice peak), after 10s split is off for 2 min. and after on for remaining time. Injection volume - 2ul. Temp. program: 30C for 2 min, after to 330C, at 30C/min and holds for 2 min. C40 comes at 13-th minute (15mx0.25x0.25 column). C40/C20 = 0.87. With longer period of initial split on, discrimination appears for the low boiling point hydrocarbons.
I think, it is not possible to use on column injector for that analysis, at least for injectors that use hot oncolumn and do not have septum purge.
Best regards,
remi
-
- Posts: 11
- Joined: Mon Dec 13, 2004 2:29 pm
what varian do you have?
-
- Posts: 11
- Joined: Mon Dec 13, 2004 2:29 pm
Dear Remirama,
I have the same system Varian CP3800.
I do C10 to C60 on it.
split ratio 1/50 after .5 min
temp. prog. start 50(2) ramp 30C/min to 350C(0)
Injector temp. 350C
detector 360C
Flow 10ml/min
used high temp column db1 (30mx 0.32id x0.1 column)
it works great!!!!!!!!have fun
I have the same system Varian CP3800.
I do C10 to C60 on it.
split ratio 1/50 after .5 min
temp. prog. start 50(2) ramp 30C/min to 350C(0)
Injector temp. 350C
detector 360C
Flow 10ml/min
used high temp column db1 (30mx 0.32id x0.1 column)
it works great!!!!!!!!have fun
-
- Posts: 18
- Joined: Sun Sep 12, 2004 5:33 pm
dear Remirama,
I don't apply split ratio an my injector is isothermal.
Best regards
carmen
I don't apply split ratio an my injector is isothermal.
Best regards
carmen
Thank you for your answer, but just to be sure two more questions. Are you turning on split after some time (~5 min)? Is injector isothermal (all the time 340 or temperature program)?
Now i'm working with such configuration: injector 40 to 340C, split is on initialy (vents out hexane, very nice peak), after 10s split is off for 2 min. and after on for remaining time. Injection volume - 2ul. Temp. program: 30C for 2 min, after to 330C, at 30C/min and holds for 2 min. C40 comes at 13-th minute (15mx0.25x0.25 column). C40/C20 = 0.87. With longer period of initial split on, discrimination appears for the low boiling point hydrocarbons.
I think, it is not possible to use on column injector for that analysis, at least for injectors that use hot oncolumn and do not have septum purge.
Best regards,
remi
Carmen
-
- Posts: 1
- Joined: Mon Mar 07, 2005 11:40 am
Dear Remirama,
try to look at DANI web site (www.danispa.it). There is an application note with a chromatogram and instrument parameter required. I use a DANI system and I'm satisfied.
try to look at DANI web site (www.danispa.it). There is an application note with a chromatogram and instrument parameter required. I use a DANI system and I'm satisfied.
8 posts
Page 1 of 1
Who is online
In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 27 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 27 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science