Acetaldehido
Acetona
Acetato de etilo
Metanol ISTD (ter amyl alcohol)
Sec-Butil Alcohol
Alcohol PropÃlico
Alcohol IsobutÃlico
3-Pentanol
Acohol n-ButÃlico
Acetato de n-Amilo
2,4Dimetil 3-Pentanol
Alcohol IsoamÃlico
Alcohol n-AmÃlico
Furfurol
Both equipments are calibrated with internal standard method. But one, the 6850 has only three calibration points and the other four calibration points. How can I be sure of the accuracy of the methods that I am ussing? I use a control sample with known quantities of each of the components calibrated and run this control sample and check for the accuracy of the method. I calculate an absolute error % between the expected amounts and the measured amounts. The % error I get is the following in each equipment:
Name| Error % 6890| Error% 6850
Acetaldehido| 49.74| -43.24
Acetona| -1.45| -16.66
Acetato de etilo| 0.89| -21.89
Metanol| -0.26| -13.73
ISTD| 0.00| 0.02
Sec-Butil Alcohol| -0.17| -7.31
Alcohol PropÃlico| 0.44| -9.00
Alcohol IsobutÃlico| -0.80| -9.29
3-Pentanol| 0.67| -17.21
Acohol n-ButÃlico| -2.33| -8.11
Acetato de n-Amilo| -2.17| -7.50
2,4Dimetil 3-Pentanol| -1.72| -10.80
Alcohol IsoamÃlico| -1.76| -7.12
Alcohol n-AmÃlico| -2.30| -6.42
Furfurol| -2.17| -9.13
So you can see that the 6850 with only three points of calibration is less exact than the 6890 using 4 points of calibration. I think that what is needed is to migrate from the three point calibration table to the four point calibration table, but the management in the laboratory says that it doesn't matter and that we don't need to upgrade to the four point table because of the time involved in making an extra point in the table. They said that is sufficient to have an % error less than 10%. I don't know what is the %error normally accepted in Gas Chromatography. Can somebody please help me with this?
Thank you for your help. Regards, Adolfo