One of the manufacturers of SPME fibers and SPME accessories is Supelco (I don't work for them

)
At the following link you can find a variety of published material for SPME and its applications (on Supelco's site):
http://www.sigmaaldrich.com/analytical- ... ction.html
For general introduction of SPME, its applications, method development and quantitation you could read in some of the published books by Janusz Pawliszyn (the inventor of SPME)
What helped me a lot when I started with SPME were some "available" web-published literature from the manufacturers. For example:
Choosing appropriate SPME fiber:
http://www.sigmaaldrich.com/etc/mediali ... p/6727.pdf
Interfacing SPME with HPLC:
http://www.sigmaaldrich.com/etc/mediali ... p/4760.pdf and
http://www.sigmaaldrich.com/etc/mediali ... p/4642.pdf
SPME theory and optimization of conditions:
http://www.sigmaaldrich.com/etc/mediali ... p/4547.pdf
References for SPME applications:
http://www.sigmaaldrich.com/etc/mediali ... p/8652.pdf
So generally you can find useful information on SPME on the web.
Regarding your present application, you have to think about obtaining:
1. Suitable SPME fiber (PDMS/DVB or PA for less polar and mid polar analytes, Carbowax-CW for very polar analytes; if your pyridine analogs are fairly nonpolar, then PDMS 100 or 30um is suitable choice for you). Remember when purchasing SPME fibers, ask for SPME fiber that are suitable for SPME-HPLC interface. Note that you'll have to choose suitable solvent for desorption of your analytes from the fiber. And for these questions consult the (information from the) manufacturer about the solvents suitable for SPME fiber desorption. Some solvents could cause swelling of the polymer fiber coating and shorten its lifetime.
2. Choose suitable extraction conditions: vial size, septa type (i prefer Teflon-PTFE coated silicone septa), agitation and temperature of sample, pH and ionic strength (salting out effect) of the sample
3. You'll have to choose whether you'll use special SPME desorption chamber connected with your LC injector or you'll desorb your fibers in to a clean vials filled with the desorption solvent. The SPME desorption chamber is better solution for better reproducibility and minimal loss of analytes
Just for information, isn't is more suitable to do SPME-GC analysis of your analytes? Are they (semi)volatile? Do they decompose with increasing of the temperature?
Hope this helps