ECD radiation leak wipe test Agilent 8890 uECD

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
When doing the wipe test on an ECD, do you just disconnect the column from the makeup gas adapter, or do you actually drop the makeup gas adapter and wipe the outside of that?

The instructions here aren't real clear, though the picture looks like it has the adapter disconnected. https://www.agilent.com/cs/library/user ... 0-3586.pdf
When doing the wipe test on an ECD, do you just disconnect the column from the makeup gas adapter, or do you actually drop the makeup gas adapter and wipe the outside of that?

The instructions here aren't real clear, though the picture looks like it has the adapter disconnected. https://www.agilent.com/cs/library/user ... 0-3586.pdf
Looks like it - the way I've done it in the past (not on uECDs though) has been to just wipe around the ferrule/nut without disconnecting the column.

CW
"Get ready, I'm hard to read like graffiti, but steady -
Science I drop is real heavy..."
2 posts Page 1 of 1

Who is online

In total there are 7 users online :: 1 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Google [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry