Hello everyone,

I am currently troubleshooting a massive air leak in my Agilent 5975 Series MSD. When checking the manual tune menu, the abundances are extremely high: >500% for nitrogen (m/z 28) and >100% for oxygen (m/z 32), relative to mass 69.

Interestingly, the system seems to pump down without any issues. The turbo pump comes up to 100% speed within 5 minutes of pumping down, and the pressure rapidly reaches e-07 Torr.

To isolate the MSD, I disconnected the chromatography column and installed a blanking nut directly on the MSD transfer line. Unfortunately, the leak persists.

The background is currently so high that I am unwilling to leave the source filaments on for any extended period. Because of this, I am unable to safely use compressed air or argon to leak-test specific connections.

Maintenance and troubleshooting attempted so far (none have resolved the issue):
> Tightening the blanking nut.
> Using an Agilent-recommended rethreading tool to remove any brass shavings from the MSD transfer threads and ensure proper alignment.
> Checking the PFTBA level (it is adequate).
> Replacing the O-rings at the PFTBA mount and the vent valve. And wiping the connecting down with ethanol.
> Cleaning and inspecting the large main MSD chamber O-ring with ethanol before pumping the system back down. (have not taken the o-ring out for a complete wipe down)

One other symptom I have noticed is that the leak is substantially worse when the transfer line heater is off --> so I am suspecting the issue might be in the transfer line weldment or a failing seal at the transfer line interface, where thermal expansion helps temporarily seal the gap.

Has anyone experienced a leak of this magnitude on a 5975 where the vacuum still comes up so quickly? I would especially appreciate any advice on what to check next or how to track down a leak this large without risking damage to the filaments.

Thank you in advance for your help!
The leaks I have experienced have either been very low and a slightly loose fitting which is fixed with a tightening of a nut or so large the chamber door doesn't stay. My first thought was that there is a minor spec on the chamber O-ring, but you mentioned cleaning it so I doubt it is there.

You mentioned that the transfer line heater off makes the leak worse, so I agree that there is something there. Could you plug the end into the MS, instead of the GC side? I am wondering if that would "fix" the leak and isolate it into the transfer line. I don't have one in front of me so I'm not sure if there is a weld somewhere else associated with it that my suggestion wouldn't help.
2 posts Page 1 of 1

Who is online

In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry