Connecting a Dionex AS-AP autosampler to an ICS 2000

Discussions about IC and related topics

2 posts Page 1 of 1
I'm trying to get the As-AP to work with the ICS 2000 under chromeleon ver 7.2. The AS appears to fill the loop (pushfill mode) but the injection valve on the ICS 2000 never switches. I have previously operated without an autosampler and used chromeleon to switch the valve to inject, that works fine. I feel like I'm missing something in the method script for the autosampler/ICS combo. I have added the AS-AP as a module and that seems good.
Any suggestions?
Hi, please go into your ICS-2000 options in the instrument configuration and ensure the valve is set to be controlled by the sampler.

See section 18.2, Figure 107 in the ICS-2000 operator manual for more details:
https://documents.thermofisher.com/TFS- ... 858-EN.pdf
Best regards,

John Guajardo
Product Manager
Thermo Fisher Scientific
2 posts Page 1 of 1

Who is online

In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry