Empower Report for SST monitoring
Hi everyone!
I am trying to create an Empower report in which I can monitor the SST parameters. I want to display a graph showing the corresponding parameter for each injection in a whole project. To do this, I am using the ‘Summary Line Plot’ report group. In principle, this works. However, if you have to process a sequence again, you have two results per injection. Is there a way to add a filter to the plot that only displays the result with the higher result ID?
Thanks in advance for your response!
Juwe
The way to do this is to set up a filter in your "Results" tab in the project so that when you select the data before plotting it, its already the maximum Result ID. Select the Edit View tab in Results tab and under Result ID cell right click, select Group By and select: To Find Results with 'MAX' ID when results are grouped by Result ID. Save the view filter and if you need to reprocess data, select only results from this filter which will be grouped by the MAX Result ID number and hence the correct data will be plotted.
In total there are 35 users online :: 0 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 35 guests
