I switched to a splitless pressure pulse inlet and after chemstation loaded method to GC, this error message appeared. When I ran the method, the error cleared. How do I check the GC to make sure the pressure pulse is being used. The display on he GC inlet showed splitless pulse pressure and the parameters of the method is identical to the GC's. Thanks.
-
- Posts: 77
- Joined: Tue Apr 22, 2008 2:11 pm
One way would be to set up the second data channel to monitor the inlet pressure. Or, to keep it simple, watch the inlet pressure reading on the instrument during an injection. You should see it increase and decrease.
Rich
Rich
2 posts
Page 1 of 1
Who is online
In total there are 77 users online :: 1 registered, 0 hidden and 76 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 76 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 76 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science