Dear experts,

I am try to automate calculation in Empower 2 using custom fields that will be locked (not editable by other users). Every thing goes Ok when the chromatogram contains ONE peak .I can add a custom field eg to calculate amout of Salicylic Acid (Peak A),
[b]Concentration_of_A= Amout * 5 / 3[/b]

The problem occurs when my chromatogram contains more than ONE peak eg , Peak A , Peak B and peak C . each peak is calculated as follows :

[b]Concentration_of_A= Amout * 5 / 3[/b]
[b]Concentration_of_B= Amout * 6 / 3[/b]
[b]Concentration_of_C= Amout * 7 / 3[/b]

It's obvious that each peak have different factors , so the same equation can not be used . I need to make a custom field that is smart enough to get the constants for each peak .

I got a simple solution by storing each constant in the CONST1,CONST2,... IN THE component table . It worked fine by I have another problem that it is editable by users. when i tried to lock it , I had to lock the entire processing method which is not feasible since each user has to edit the processing method (eg changing Threshold , Retension times ,...etc and otehr parameters that changes between different runs)

I hope the situation is clear now . for any further inquires , please feel free to ask .

and thank you in advance for your help .
Dear All ,

Thank you all in advance for your help .

You can make use of CCompRef1, 2, 3 ... Bu this is also editable (altough not so clearly visible factors).

Another way is to setup your custom fields and to calculate each factor for each peak eg:

for peak A:
Amount * 5 / 3
Amount * 6 / 3
Amount * 7 / 3

the same for peak B and C.

Then in your report make different tables for peak A, B and C within the report and show only the appropriate custom field.
You can also lock your report.

Or you can make use of the EQ function for example EQ(Name,"Peak A")*amount*5/3 + EQ(Name,"Peak B")*amount * 6/3 .....

Kind regards

Ace
3 posts Page 1 of 1

Who is online

In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry