I have been trying to adapt an IC-ICP-MS/MS method for Sb speciation using an Dionex AS22 Column and unsurprisingly I have been having problems with low resolution between SbV and III.
The base eluent i have been using is 1.5 mM Phthalic acid and 10 mM Na2EDTA in 2:98 MeOH:H2O with the pH adjusted to 4 using Ammonia and HCl. I tried to add 4 mM Oxalic acid (from 0.2M solution in H2O, so 2% of the Oxalic acid solution and 98% of the base eluent) with the LC quaternary pump and the chromatograms looked absolutely fantastic, but after an hour or two the backpressure started to rise rapidly leading to the LC shutting down. I was able to slowly flush the system with 2% MeOH in H2O and get the pressure back down. I am using Type 1 water and Analytical grade or better chemicals.
My guess is that some kind of poorly soluble oxalate is forming in the system and clogging it?
The LC system i am using is an Agilent 1260 Infinity II Bio-Inert so my reckoning is that the eluent path materials shouldn't be a problem.
I also tried with the oxalic acid solution also being adjusted to pH 4 but the chromatograms were useless.
So i guess my question is if anybody has any ideas on how to prevent this/stabilize the eluent solution since i would very much like to achieve a reliable way to get the chromatograms that i got with the oxalic acid.