Agilent TQ 6495C changing responses
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
Hi! For a week or so our 6495C have weird behavior. After tuning, which pass great, in jus a few hours responses for ions falls approx. 10 times, and even faster decreases when there is changing methods between runs (after pesticides in food, doing ochratoxin A, then returning to pesticides.) Then, it somehow regenerates overnight, sometimes, but it always running good just after tuning. From the service, they had idea about quads drivers, but it was in good shape, I think they concluded that, so they abandoned everything about hardware. About that, I must mention that they did PM with a lot of hardware preventively changed, and cleaned optics and quads, it was barely month ago, about in January. So we are now in magical loop, it runs good for a short time, then it goes crazy, and after tuning the loop goes again. Anyone have ideas about what happening?
-
- Posts: 11
- Joined: Thu Jan 05, 2017 2:27 pm
I don't know if you already solved the issue? Our Agilent 6460 had very similar symptoms. When tuning, everything was fine, afterwards, the responses were like a rollercoaster. The quad drives were replaced, they even cleaned the first quadrupole rods and changed the x->y orientation. There was no improvement. Still inconsistent results. In the end, it came out it was the ion source that was the culprit. Two things affected the nebulizer performance:
- the yellow plastic ring around the nebulizer assembly was cracked and leaking nitrogen,
- the electrospray nebulizer locking screw on top of the AJS ESI was not tightened enough, so it has a little play, probably also leaking nitrogen
When both issues were resolved, the responses were a lot more consistent.
- the yellow plastic ring around the nebulizer assembly was cracked and leaking nitrogen,
- the electrospray nebulizer locking screw on top of the AJS ESI was not tightened enough, so it has a little play, probably also leaking nitrogen
When both issues were resolved, the responses were a lot more consistent.
2 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 1 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 2 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 2 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science