Simple Standard for DB-5 MS column performance check

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello!

Anyone could recommend me a simple standard/ mixture to check the performance of a DB-5MS column (using MS detector)?
I am thinking of a mixture of analytes/ solvents, nothing too expensive or that requires certification.

Thanks in advance :)
Whip up something like this.

https://www.agilent.com/store/en_US/Pro ... 0/200-0310

Whenever I've purchased a new GCMS system from Agilent, they install a 5-phase column and use this test mix to ensure it's good.
Consider using the Grob Test Mix - I’ve used it to monitor performance and it has a lot of historical data on what it should look like. It will also almost certainly stay available in the future.

Available from Merck, Restek and the other usual suspects.

https://www.sigmaaldrich.com/US/en/product/sial/86501
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry