IC retention time shift.
Discussions about IC and related topics
Hello everyone,
I’ve observed a significant shift in the retention time for ammonia, which has increased from a consistent 4 minutes to approximately 12 minutes. The method had been well established with a retention time around 4 minutes, and this change occurred unexpectedly over the weekend.
To troubleshoot, we have already taken the following steps:
Replaced the column
Flushed the suppressor
Flushed the system with deionized water
Cleaned the check valve
Inspected all connections
Primed the system to remove any air
Despite these efforts, the issue persists. Any insights or suggestions you can offer would be greatly appreciated.
What about the retention times of other cations?
What about the flow rate accuracy? Did you verify it by measuring with a volumetric flask and a stopwatch? What about the pressure ripple?
How do you prepare the eluent? Do you use electrogeneration?
Any clicking noise from the motor? That is a sign that it is failing and it can shifts RTs. What about the column temperature? If that changes the RT would shift.
In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)