There are three cell windows (lenses) in the path from the lamps to the detector: Coupling Lens Assembly, Source Lens (Achromat) Assembly and Cell Support Assembly. I do not replace the coupling lens assembly one as it lies between the two lamps and I do not run Vis wavelengths. I have had great success swapping out the Spectro window P/N: 79880-28111 in the other two assemblies for noise issues when I determined the lamps and flow cell are good. These Spectro windows/lenses develop what looks to me a white colored film/burn on them over time. This swap out has fixed my noise issues on 19 of 20 DAD and MWD's over the years when new lamps and flow cell still failed for noise. The one time it still failed after swapping these I swapped out the entire optics bench from a parts system and then that solved my noise issue.
Caution to do these swaps you need to pull the optics bench out, pull off the Holmium filter holder (can be a pain in itself), pull the assemblies out (can be a pain) and then disassemble the cell assemblies (can be another pain) then put it all back together. You probably could purchase the entire cell assemblies for much more $$$. I got this procedure from an old 1100 DAD/MWD detector manual, back in the day. Agilent took this procedure out from newer versions of the 1100 and 1200 manuals I presume because people were totally breaking their detectors trying to do it. I haven't had a noise issue on my 1260 detectors yet so I never dived in to see if it is a similar design. I do not have the manual anymore, I just learned how to do it back in the day.