I'll give a bit of context to answer any potential questions right away.
Our production facility uses commercially bought denatured Ethanol in it's manufacturing process. The Quality Control lab where I work is charged with analyzing the purity of this raw material before use in the process that will, eventually, lead to much needed medicines.
To this end, we use an in-house developed method and a 7820A GC-FID system.
For the last 8 years, no problems were reported, other than the occasional failed run. Failures that were easily remedied by maintenance (changing the liner and/or septum, cleaning the column regularly, preventive 6-monthly maintenance by the supplier).
Our analyte is pure ethanol (around 93%) denatured with Methyl-Ethyl-Ketone (around 2%). Injection by the autosampler was 2µl, on a 10µl Gold Standard syringe. Carrier gas is H2.
The way our internal System Suitability Test is set up, the calibration standard is injected 4 times : 1 time as a "dummy run", 3 times as "actual" runs. The ratio of Analyte/internal standard is calculated; if the RSD is greater than 1%, the run is failed.
Around November 2024, we started getting poor repeatability on the three calibration standard injections. Numerous runs were suddenly "invalid" due to a high RSD. It started relatively benign (around 1.6% RSD) but got to 4.5%.
The invalids popped up across the lab : all the technicians that have performed this analysis for at least 3 years without problems suddenly encountered this drop in repeatability.
Every part that could be changed on this machine was changed (by qualified technicians/engineers from the supplier).
The method did not change.
The technicians follow the Standard Operating Procedure scrupulously.
Still repeatability problems.
It got to the point where we would install a new needle (pre-cleaned with acetone before use), get one or two valid results out of it, and the repeatability of the injections would drop again.
We installed the (slightly adapted) method to a brand new GC (8860 GC-system). First testing and validation of the method went well, but now we use it in routine, we encounter the same problems (RSD of 11.5% sometimes).
Interestingly, as soon as we switch the syringe for a new one, it works, but again, for one or two runs (which typically consist of 7 injections each). When inspecting a used syringe, it is as if the syringe does not aspirate the fluid it is supposed to inject. The latest syringe (still a Gold Standard 10µl syringe) worked perfectly well on Monday 03 MAR 2025, faulted Wednesday 05 MAR 2025.
I tried a manual cleaning of the syringe, and noticed that when I carefully move the piston, liquid does get into the syringe but a multitude of small bubbles are formed. The same result happens at different speeds of moving the piston.
Other, older, syringes, seem to have lost the capacity for aspiration altogether : nothing gets aspirated in the graduated part.
It is clear our problem with repeatability is due to a faulty and unequal injection of the mixture, so logically the fault lies with the syringe. But since we change it regularly (twice a week!) and the problem is recurring, there must be something else going on.
Did anybody encounter similar problems? We (= my colleagues and myself) do not know what the root cause could be. The technicians/engineers of the supplier are equally stumped. Routine analysis has become a headache and if frustration was marketable, I would be a millionaire by now...
Any ideas? Tips? Pointers?
Thank you for reading this long winded post