Trace Ultra gc oven temperature sensor malfunction
Hi, very new to GC FID and the Trace gc Ultra device. On powering it up it gives the hardware shutdown screen because it thinks it is at 632oC. Does anyone know how to replace the oven temperature sensor (assuming that’s the problem). Apologies for such a basic question, Jim
First , you may check the sensor if it is really faulty and needs replacement.
You need an ohmmeter to check it.
Unplug the instrument ,remove the right hand side panel of the GC , there is a hex screw at the back side of the panel , loosen it , slide the panel to the back , tilt and remove.
You will see multiple sensor connectors on the electronic board , locate the '' oven '' sensor connector , it has two wires , check the resistance between the wires , it must be around 105 -110 ohms when the oven is cold.
If it is open circuit ( no resistance reading ) or short ( zero ohm reading ) , it is faulty.
For replacement of the sensor , you may google '' Thermo Trace GC service manual '' and you will get the manual , there is full explanation of sensor change procedure on the manual.
Good Luck !
Good Luck !
Hi Uzman,
I really appreciate the quick response to my question. I have the manual and check the resistance of it tomorrow and report back. Fingers crossed that is all that is wrong. Many thanks, Jim
Hi Uzman,
I checked the resistance of the oven temperature sensor as you suggested and it read 103.6 ohms. On reassembly it now seems to be starting up robustly and reading correctly without error. Looks like it might have been a bit of a dry contact and the investigation with confirmation that the sensor was not faulty sorted the issue. Very many thanks for your help - greatly appreciated. Jim
You wellcome Jim , good to hear that it worked out !
In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 1 guest