I've been experiencing problems with a GC shutting down recently as it thinks there's a leak at the inlet. I came in to find it alarming with hydrogen safety shutdown on the display. Understandable given we use hydrogen as the carrier gas and it thinks there's a leak.
Initially I went through the usual things to resolve this e.g. replacing septum, making sure fittings were tight enough and upon resetting the GC the initial settings were all reached and the problem appeared to be solved.
However I then got a call saying it had happened again after an injection had been made. The timing was undetermined as the injection had been made and the operator left the lab. They came back to find it alarming.
Having reset the system the issue again cleared so it doesn't seem like there's actually a leak at the inlet as surely it wouldn't be able to get up to pressure again if a leak had occurred. Checking the actuals confirmed steady pressure and flow at initial starting conditions. The method is set to constant flow.
I made an injection myself and watched to see what was happening. It seems the problems begin to kick in as soon as the ramp starts with the instrument beginning to beep. The pressure at the inlet climbs, as you'd expect, but the flow begins to drop steadily from set point of 1.07. The instrument then shut down about a minute after reaching final temperature at the start of final hold (240degC) and the flow was c. 0.6. Again, hydrogen safety shutdown on the display.
Does this seem like a settings issue in the method (I've checked the method and constant flow is selected) or something more sinister?
Any advice would be much appreciated.