MS ESI ionization of methoxybenzenes in normal phase LC-MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi all,

I am running in some great trouble to see the [M+H]+ peak on positive ESI mode. I am analysing methoxybenzenes like 1,3-dimethoxybenzene (138.07 m/z) . The LC-MS is a normal phase system where THF is used as eluent.

Two main problem arise;
1) Even when infusing directly the 1,3-dimethoxybenzene in methanol in the MS, I get an unknown peak (+304 m/z) that I can not correlate to any adduct.
2) The system should run on tetrahydrofuran (1 mL/min) and methanol (0.5 mL/min) is added just before the MS, to improve ionisation. Unfortunately, THF gives very poor ionisation, to the point no increase in the TIC signal is detected when 1,3-dimethoxybenzene is injected in the system (using a bypass column). The peak is well visible in the UV, but no signal results in MS.

I really need to find reliable parameters to have at least some signal for an adduct I can find with all my methoxybenzenes. At some point I observed the peak [M+Na]+ but it is no longer the case for some reason.
I tried to change the ionisation solvent (0.5 mL/min) to the following solvent list, but nothing worked:
MeOH
MeCN
0.1% HCOOH MeOH
0.1% HCOOH MeCN
10 mmol/L NH4HCOO in MeOH
1 mmol/L NH4HCOO in MeCN
1 mmol/L AgNO3 in MeOH
1 mmol/L AcONa in MeOH

What to do now? It seems that most often people see at least some [M+H]+ peak, but in my case it is non existent... What to optimise or try first?
Do not hesitate if you need any more info!

I used the following parameters:
cone voltage: 4.5 kV
dessolvation line voltage: 0V
heat block temperature: 400°C
dessolvation line temperature: 250°C
Hi,

I have no experience with these particular molecules, but I would intuitively be tempted to favour APCI over ESI, and RP-LC over NP-LC. Is it something you could consider testing or are you stuck with this set up? Also, if you have a GC-MS in the lab, that may be worth a try.
As a follow up on my question, I could manage to obtain some ionization with isopropanol as ionization solvent. Although it is not so reliable. Indeed a reverse phase system would be much better for MS signal but in my case it is not possible...
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry