- Posts: 2
- Joined: Tue Feb 04, 2025 3:14 pm
- Location: Italia
We are developing an analysis method to determine the following analytes: Chloramphenicol, 2,4-di-tert-butylphenol, Palmitic acid, Propylparaben and Stearic acid.
We are developing the method in negative and apart from Chloramphenicol, even injecting a 1 ppm standard, we do not see the peaks in chromatography but only extracting the ions and they are very low.
We are using an Agilent 1260 Infinity II combined with Agilent 6530 LC/Q-TOF.
as a source we have a Dual AJS ESI.
the QTOF parameters are:
Gas temp:250
Sheat Gas Temp: 250
V Cap: 3500
Fragmentor: 150
The column is a Poroshell 120 EC-C18 3x150 mm 2.7µL.
T column : 45 C
Mobile Phase-A ACN + 2 drops of ammonia
Mobile Phase H2O + 2 drops of ammonia
The gradient we use is:
0 min 10% Mobile Phase-A
1 min 10% Mobile Phase-A
5 min 90% Mobile Phase-A
20 min 90% Mobile Phase-A
25 min 10 % Mobile Phase-A
35 min 10% Mobile Phase-A
Flow:0.4 ml/min
do you have any suggestions or ideas to improve the ionization?