Good morning. Today I changed the big trap helium in the Agilent GC(7890A)/MS(5975C), but as before, I continue to have a high nitrogen (47.8%) and oxygen (12.7%) value on tune. I tried change the helium tank, the septum and the liner but nothing. I think it's an air leak, What can I do?
I think it's an air leak
I think so too.

I'd block off the MSD at its inlet in the GC oven, and see what that shows. That should isolate whether such leak is in the MSD or the GC.
Same idea, slightly different. Disconnect the column at the inlet (cold) and plug it with a septa. Means you don't have to vent the MS. Then spray with dust off or dry ice or isopropanol and look for appropriate ions.

Best regards,

AICMM
Same idea, slightly different. Disconnect the column at the inlet (cold) and plug it with a septa. Means you don't have to vent the MS. Then spray with dust off or dry ice or isopropanol and look for appropriate ions.

Best regards,

AICMM
Good idea.
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry