- Posts: 506
- Joined: Fri Jun 02, 2017 11:10 pm
- Location: Western Colorado
are:
1,2-DCA 0.38-5 ug/L
Benzene 5 ug/L
EDB 0.02- 0.05 ug/L"
By running 8260 in SIM/Scan and using SIM to quatitate I have these MDL's respectively for water matrix...
1,2-DCA 0.18 ppb or by Scan mode without SIM... 0.59 ppb
Benzene 0.20 ppb or by Scan mode without SIM... 0.48 ppb
1,2-EDB 0.19 ppb or by Scan mode without SIM... 0.49 ppb
Am I right in thinking I cannot get any lower for 1,2-EDB without going with an ECD detector method?
My current SIM/Scan 8260 method cal curve is from 20 ppb to 480 ppb. I used a 50X dilution of the CCV standards to run a 0.8 ppb and a 4.0 ppb check standard each day I ran 8260. I used student's T test on the averages and standard deviations of of each level taken from 16 different runs to get the MDLs' I cite above (c.f. CFR 136B how to determine an MDL).