Yes, I really hate how the reset switch is designed on the Xevos.
How about through Console | System | Troubleshoot | Scan Instruments
Does it give IP addresses for the MS and all modules? Perhaps a cable is loose or the IP address was corrupted somehow?
I believe I had a similar problem in the past. Not only did we have to power cycle, but also unplug the MS power cord for 5-10 minutes to give time for the capacitors to discharge.
And for others who might be stumbling across this topic, there are several ways to reboot the epc. Through the pinhole on the front of the mass spec; through Tune | Setup |Communications Status; and through Hyperterminal (or equivalent) reboot command.
Much of this may apply as well:
https://support.waters.com/KB_Inst/Mass ... n_MassLynx
Hi,
Our XevoTQD hooked up to Acquity H-Class UPLC is not communicaitng with Masslynx. We can't conduct EPC reset through the hole since that button is not responding (not making any beep noise). We can't do EPC reset through command lines since it is not pinging EPC (MS)...
Since soft-reboot wasn't possible we tried power cycle (venting and recycle) but it is still not connecting to masslynx or showing up on DHCP server.
Please let me know what I can do. Thanks