PPC SHUTDOWN ERROR

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Hello, Could someone please tell me what to do if Clarus 580 shows the following error messages: CAR 1 PPC SHUTDOWN and SPL1 PPC SHUTDOWN. Thank you
Hello

Step 1 is to check that carrier gas is getting to the inlet. Is the pressure setpoint the same as the pressure actual?
Hello, thank you for your reply.
When I lauched my analysis, the actual pressure and setpoint pressure were same. But, midway it started beeping, and the pressure went to zero and is showed SPL1 PPC SHUTDOWN on the GC interface.
If you provide detailed info, you'll probably get better answers here.

Watch the pressure during a run. The setpoint might increase along with the temperature, and maybe the supply pressure isn't high enough to keep up with it. The solution would be to make sure the supply pressure from your regulator is high enough, and (as is the case 90% of the time with GC) to fix any leaks.
Hello,
Thank you so much for your response. The problem is resolved, turns out the pressure regulator was not working properly and as a result the carrier gas pressure was not reaching the setpoint.
5 posts Page 1 of 1

Who is online

In total there are 328 users online :: 1 registered, 0 hidden and 327 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 327 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry