I've done some back of the envelope calculations and I'm not seeing much difference in the assay result, so I think it doesn't matter too much for that aspect. We're just trying to convince our QA folks that we can calculate the required resolution using a minor component that is below LOQ.
The two components under discussion are not isomers. They are homologues. The only reason to strive to separate them and to have their resolution as a part of SST is the necessity to control their ratio as described in USP (in other words, to control the fraction of the minor component; this implies that the check of the signal-to-noise ratio must be a part of SST as well). The standard to be used for SST must be suitable for this purpose. It must contain sufficient amounts of the substances needed for the resolution check. Your current standard contains insufficient amount of the minor component rendering the standard unsuitable.
This questions the competence of the laboratory choosing the standard, the competence of the USP authors who don't specify the disregard limit for the minor component in their method, and the competence of the standard manufacturer that does not take into account all the aspects of the intended use of the standard. You may try to convince your QA dept. to calculate Rs with the minor component below LOQ and above LOD. However, when you come across with a new batch of the standard with the minor component below your LOD, you will have to search for another standard.
Another option will be to try to lower your method's LOQ and LOD, e.g. by some minor changes such as decreasing the data acquisition rate or using VWD instead of DAD (both variants lower the noise). You can also increase the injection volume; the current USP chapter 621 Chromatography permits this (if linearity and the other SST are still satisfactory).