Less sensitive wavelength to avoid dilutions?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
I know that typically, we choose the wavelength of max absorbance for analysis when using UV detectors. I am wondering how common it is to choose another wavelength with lower sensitivity to avoid having to perform additional dilution.

I am developing a new method and would like to use this approach, but wondering if there could be any downside that I am not considering assuming that sensitivity is not a problem and there are no interfering peaks.

I guess if you choose a part of the spectrum where the slope is very steep and your detector is not very stable, that could result in increased variability. Any other potential issues?
You can use a lower maximum or use a minimum (if they exist) in the spectrum. It is OK.
An option is to reduce the injection volume.
I should have mentioned, there are two analytes with very different Uv spectra. The original method used a wavelength where one has much higher response than the other so different dilutions were required. So in this case, decreasing injection volume won't help. I am hoping to find a wavelength where both can be in linear range with a single dilution.
I should have mentioned, there are two analytes with very different Uv spectra. The original method used a wavelength where one has much higher response than the other so different dilutions were required. So in this case, decreasing injection volume won't help. I am hoping to find a wavelength where both can be in linear range with a single dilution.
Can't you time-program a wavelength change? I'm retired but our detectors "back in the day" could do that.
Agreeing w/ ^ - so long as there's enough resolution between the peaks, a wavelength switch is frequently a good way to assay things with wildly different absorbtivities.

Do have a look at the spectrum... a local maximum is an ideal target for that non-max absorbance wavelength. The thing to avoid is a region of the spectrum where absorbance is changing rapidly.
Thanks,
DR
Image
5 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry