Typical HiVac pressure for 5977B?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello all,

I'm curious what the group sees as typical and good pressure on HiVac gauge on the 5977B. We realize it depends on column flow and there is a table in the manual.

Our system is a Headspace GC/MS (Agilent 8697/8860/5977B). When running 1.2 mL/min through the column we are reading 1.03 x 10^-5 after a couple days of pump down time. Instrument is 3 years old, light usage.

Is this typical? Curious what you all see. The manual says we should be within 30% of 7.33 x 10^-6 at this flow rate. We are just slightly higher than that (40% higher).

Thanks!
More importantly, what does your air/water check look like? If this is good, I would expect a slightly better vacuum, say 5 to 8 x10e^-05. How old is the rough pump oil?
Regards,

Christian
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry