I'm new to writing on this forum, so I hope I'm doing this right! I’m looking for some help to understand a phenomenon I encountered recently.
I’m designing the Qualification / Maintenance tests for an Agilent RI HPLC. During a dry run of all the tests we usually perform with our DAD HPLCs, everything passed except for the injector linearity test. This test involves using the same concentration sample with different injection volumes.
The response I observe is non-linear, more like a logarithmic curve or an asymptote. To rule out detector overload, I lowered the concentration of the standard to a third, but the non-linear behavior persisted.
I get that this test is pretty much worthless since all our methods are ran at constant inj volume. I only did it because we do it for the DAD HPLCs, and my organization tends to stick with “we’ve always done it this way,” so any change will require some convincing.
I did a bit of online research, but I couldn’t find a solid explanation for why this happens. It seems to be tied to the fundamental operation of the RI detector, but I don’t quite get why.
Does anyone have a good explanation for this?
Thank you very much!