1. Autotune only? No manual tune? I talked to the app chemist at Agilent and he says I should be able to see manual tune. I do not seem to have any options or tabs that will point me to the manual tune.
2. In the ESI negative, in order to meet MRL of 1 ppt for some of the hormones, I am at gain 10 which is the highest gain allowed while using 20 uL injection. Is this normal?
3. While using polarity switching, with gain 0.2 for ESI(+) and 10 for ESI(-), I am not able to produce decent curves, all are convex. When I split the method to 2 separate ones, the ESI(+) curves look great, ESI(-) not so much.
I sent data to Agilent this week but the app chemist I talked to really didn’t have any solutions except "clean the capillary and retune”, so I will be talking to someone else next week. In the meantime any chemist out there who has seen similar issues and may have some answers and can help?
Thank you!