Thanks for replying. Years ago there was s thread on whether you should be allowed to manually interrogate. You should.
https://www.chromforum.org/viewtopic.php?t=10304
I like these 5 rules;
Rule 1: The main function of a CDS is not to correct your poor chromatography.
This places greater emphasis on the development of robust chromatographic procedures so that the factors involved in the separation are known and controlled adequately. Whenever possible, separations should be developed such that automatic integration is the norm not the exception. Management need to understand that adequate time must be given to method development and validation. This is especially true for pharmacopoeial methods that never work as written.
Rule 2: Never use default integration parameters, always configure specific integration for each method.
Without exception, peak integration and result processing must be defined and validated for each method so that all peak windows and names are defined and if necessary any system peaks are identified. Using a default or generic method results in excessive need for manual integration to name and calculate peaks.
Rule 3: Always use automatic integration as a first option and control manual integration practices.
Remember that the use of manual integration is a regulatory concern and use needs to be scientifically sound. Also be aware that, as discussed earlier, manual integration slows down a process, so see Rule 1 to get the right method depending on the sample matrix and peaks of interest.
Rule 4: Understand how the CDS works and how the numbers are generated.
This requires basic training in the principles of peak integration and how a CDS works. The problem is that with mergers, acquisitions, and encouraging experienced analysts to retire and employ younger workers, skills are being eroded and a CDS can be looked at as a black box that always gives the right answers.
Rule 5: Use your brain-think.
This rule is sometimes difficult to follow but follows on from Rule 4. You can have what appears to be a perfect separation and peak integration, but look at peak start and end placement-do they look right? Use the zoom and overlay functions of the CDS to see if standards and samples have the right peak shape. The analyst is responsible for executing applicable procedures correctly, which includes correct peak integration. The reviewer, however, also has a role to ensure that all integration (whether automated, optimized, or manually placed) follows the method guidance for placing baselines as the SOP describes, especially when the representative area for unresolved peaks are being estimated. Significant peak area manipulation should be easily noticed by an experienced reviewer.
https://www.chromatographyonline.com/vi ... -draw-line
4 and 5 weren't listed t in the article you attached. # 4 is a big problem. People want to treat the CDS as a big pH meter and just accept the answer that pops out without thinking - see rule #5. The lab missed a proficiency several years ago when a baseline problem interfered with a Nitrite peak. After that I got some things written into the SOP about magnifying the chromatogram and looking closely at the baseline. But years later the woman who made the mistake because a supervisor and wanted to undo that to simplify things. She and the other analysts didn't understand how to program the instrument's quantitative method for consistent integration either. I'm still the only one in the lab who seems to know how to do it. One woman just wanted to raise the reporting limits for things, especially fluoride, so the small baseline problems and co-elections don't need to be dealt with.
I attended a TNI training a couple of years ago and it said to maintain the original data that wasn't manually integrated. The rule I am telling people to go by and am trying to get written into the SOP is to make a copy of a file and manually integrate in the copy. Then save the copied file with an M post-fix on it so it is easy to identify as a batch with a manual integration. It's happening 2-3 times a month. Unless someone has a better idea I am going by a 10% rule. I don't know if an auditor will even bother to look at this. They have never looked at our manual integrations in the 13 years I have been at the lab - that's 7 audits. I remember one time the head auditor asked if we ever had to manually integrate and I said yes, but not often. He never looked at any. I am pretty meticulous about it. I print magnified images of the calibrator peaks and the samples peaks before and after manual integration showing how they match how the calibrators are integrated. I also note that the problem cannot be fixed by changing the quantitative method.
My lab has an SOP for manual integration but it was originally written for the organic section because until 10 years ago they are the only ones who performed chromatography. Then we starched using an IC for anions. I have complained about the integration SOP and asked that it be updated, but nothing happens ...So I am pushing for more specifics in the SOP for anions that I deal with.