Change of carrier gas

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Dear All,
This particular question is being raised with respect to Indian Pharmacopoeia. I did not find any suggestions on whether it is allowed to change the carrier gas (I want to use helium instead of nitrogen which is specified in pharmacopoeia). I am analyzing ethanol (95%) and would like to improve resolution in system suitability between methanol and acetaldehyde (NLT 1.5). With nitrogen, I am getting resolution of 1.0 with nitrogen.
If I change the gas to helium, do I need to perform method validation or can proceed with method verification?
Please help and suggest.

Best regards,

Aniket
Yes, you will need to demonstrate equivalency after figuring out your new flow rates etc. This is a significant undertaking - make sure you'd rather do the work than just replace a column or repack (if it's a packed column method).

That said, if you're going to switch, switch to Hydrogen! You'll save $$$ and possibly improve your resolution.
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry