- Posts: 78
- Joined: Tue Sep 13, 2022 1:09 pm
The system has some peculiar vacuum pump behavior.
When shutting down/venting the MS from the MassHunter software, the turbo pump slows, but the rough pump never shuts down until we power off the MSD.
Shouldn't the rough pump shut down automatically once the turbopump is below a certain speed and vent valve is opened?
On power-up, the rough pump and turbopump both start immediately as soon as the power switch is engaged. This is before the MSD has booted or even connected to the MassHunter computer. The MassHunter computer then shows Turbopump struggling at 3% and the pump down process is already underway when the MSD finally connects to MassHunter (without issuing the pump down command). In a minute or two enough vacuum develops holding the door closed that the turbopump accelerates.
Is this normal vacuum pump sequencing behavior for the 5977b? Doesn't the turbopump normally wait for a command or minimum vacuum or rough pump run time before starting?