Hello!

I would like to ask, can a degradation impurity be unspecified? In a monograph can an unspecified (but listed) impurity be degradation impurity?

Thank you in advance
Yes, a degradation impurity can be unspecified. The advantage of a specified impurity is a higher level allowed in the drug substance or product. Or much lower in the case of mutagenic ones. However, this needs to be justified (toxicological concerns). Obviously, the identification of the impurity to be specified always is required.
Most of the impurities monographed are intermediates of the synthesis and are related to a specific synthesis route. Sometimes degradants are monographed. All impurities monographed can be considered as known.
Also check ICH guideline Q3A/B and Q6A: https://ich.org/page/quality-guidelines
2 posts Page 1 of 1

Who is online

In total there are 241 users online :: 1 registered, 0 hidden and 240 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Google [Bot] and 240 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry