Separation of compounds with close hydrophobicity
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
I'm looking to hear ideas to separate two similar compounds that differ only with one containing a single Cl and the other a single Br atom. Can they be separated?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Get into the lab, and try some stuff !!! I had to do similar many times when I was working.I'm looking to hear ideas to separate two similar compounds that differ only with one containing a single Cl and the other a single Br atom. Can they be separated?
Make up standard solutions of each individually, see if you can come up with ways to see a peak in each solution. Then mix portions of those two and see if there's separation.
If you don't get separation in a mix, vary one thing at a time. For example, was easy to program in a lower column temperature, and a higher column temperature, and see if the separation got better, worse, or stayed the same. That's the start of a mini Design of Experiments.
Do similar changing columns, mobile phase, trying a gradient, etc. When I had similar situation, I found that I could achieve separation by adding a few %THF to the mobile phase; another way that worked there was to use what was new then (intrinsically base-deactivated column) which actually reversed the order of elution and made quantitation much easier. Some column suppliers can supply technical help but they'd need to know the molecules, and won't happen fast; but they would love to have you keep buying their column regularly !
If you can, post names/structures of your two molecules.
-
- Posts: 11
- Joined: Wed Feb 01, 2023 2:54 am
Hard to say if they can be separated, can you say what the compounds are and what methods (column, mobile phase) that you have tried so far?
-
- Posts: 70
- Joined: Fri Jan 02, 2009 12:53 pm
a PFP phase would be a good starting point, or pentabromobenzyl phase
note, you can get very different selectivity depending on whether you use methanol- or acetonitrile-based mobile phase
note, you can get very different selectivity depending on whether you use methanol- or acetonitrile-based mobile phase
4 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science