what is the difference between column equilibrat and column condition?

Equilibrate=Flow at initial conditions
Condition=Run the gradient

In my experience, condition usually refers to something that is done only once at the beginning of a procedure or set of samples to prepare a piece of equipment like a chromatographic column for use. Equilibration usually refers to a step that is performed repeatedly in between multiple runs in a set.

I doubt that either term is rigorously defined anywhere, so usage will vary. In any given case, the interpretation should be obvious from the context.

I'd agree with the previous posters that (usually) "equilibration" is something that's done for every run, and "conditioning" is done less frequently
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

I’d agree with AA (i.e. equilibration is an isocratic holdup, whilst conditioning of the column is exposing the column to the changes that will occur during the planed separation/elution – typically the gradient as programmed.
I also agree with Tom that no universal definition is established – as yet. But if one uses Empower, for instance, choosing one of these two functions will cause the execution of the actions as described above.
It should also be mentioned that if the instrument method is isocratic, isotherm etc. (i.e. no timed events) choosing the function “condition columnâ€
Learn Innovate and Share

Dancho Dikov
I agree. The reason i asked is that i noticed choosing equilibrate or condition in isocratic mode didn't make any difference.

Danko and all,

Just 1 point of note (of interest?) when using Empower. Selecting "condition column" runs the gradient table but does not run the pump event table. Also, neither equilibrate or condition collect any data. If you want to run the gradient table and events tables you choose "Inject Immediate Samples", this mode collects data, runs the tables but does not do a physical injection. Very useful for diagnosing pump and mobile phase problems.

I use Waters/Empower 2 – This is how I understand it:

Condition: To create the “conditionsâ€
Kind Regards,
Jade Barker

[quote="Diagnostic.Lab"]I use Waters/Empower 2 – This is how I understand it:

Condition: To create the “conditionsâ€
9 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry