Is there a deeper issue at hand with my detector or something possibly that could cause this?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
To me it sounds like some concentration-dependent process is starting to occur somewhere in the system, which may just be inherent to your compound or may be exacerbated by your conditions. As James said, can you describe more details as to what conditions you're using and the analyte you're targeting? It could also just be a bad injection as miller mentioned.Running a standard curve on HPLC and my linearity is coming out poor (0.9772) due to the highest concentration being lower than what it should be to make the line fit properly for my data points. I've made multiple new standard solutions to ensure it is not error in pipetting etc.
Is there a deeper issue at hand with my detector or something possibly that could cause this?
Can you post your calibration graph? Perhaps a linear fit is not best, maybe your concentration range is quadratic?Running a standard curve on HPLC and my linearity is coming out poor (0.9772) due to the highest concentration being lower than what it should be to make the line fit properly for my data points. I've made multiple new standard solutions to ensure it is not error in pipetting etc.
Is there a deeper issue at hand with my detector or something possibly that could cause this?

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.