Dan » Mon Jan 31, 2005 4:40 am
Kaoula,
Sorry, I made a mistake in what you meant about the peak areas with the detectors. Now, that I understand it , then yes there is a problem.
To help troubleshoot your problem, some more information would be useful. Uwe makes a good point about checking the operation of the detectors. Are the PDA and UV operating okay? Are the parameters for bandwidth, time constant, etc. set properly?
Tom also makes a good point that the way you are doing the calibration and quantitation need to be checked. The results you gave in your second message may indicate that it is not a detector problem. Can you give more information on the calibration/quantitation procedures?
Lastly, you asked about methods development using the PDA and routine (QC) analysis with UV. Yes, you can do this. Using the PDA to do the methods development is very common. There are many benefits to this. In the pharmaceutical labs were I have worked, we usually use the PDA for the development stage. However, we also run development analyses with the UV. We want to test the method as best as possible before we transfer methods to the QC group. The PDA greatly helps the methods development, but we usually use UV for the routine work.
Regards,
Dan