Waters Alliance 2790 no peaks?!

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Using an Alliance 2790 in conjunction with a Waters LCT Mass spectrometer.

The syringe seems fine, sample is being withdrawn from the vials. Mobile phase is present, and reaching the MS ok. I've bypassed the column completely so it isn't a column issue, and I can't find any leaks. However I get completely flat chromatagrams containing only chemical noise from the MP.

The MS part seems fine, however I'm not seeing any sample peaks. I've checked the needle and the injection port. Both seem unblocked and correct. I'll try the injection valve tomorrow. Any other possibilities?

Thanks for any suggestions.

Is that a standard sample that you know will produce certain peaks?

Yes.

Have your tried direct infusion in your MS to see if your MS is working OK?

Yes, the MS is fine.

Fixed.

There was glass - froma broken vial insert? - in the block between the needle port seat, and the needle. Not obvious to see!
6 posts Page 1 of 1

Who is online

In total there are 25 users online :: 2 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry