By cytw98 on Tuesday, August 17, 2004 - 11:59 am:
Does anyone know whether that Alltech's ELSD 2000ES or ELSD 2000 is compatible with Waters Alliance systems?
Thanks!
By A on Tuesday, August 17, 2004 - 12:38 pm:
This depends what you mean by compatible. Control of the Alltech would be through its front panel and data would have to be passed into an A/D converter and then into your data system.
By cytw98 on Wednesday, August 18, 2004 - 05:39 am:
Does that mean that the mobile phase gradient and flow rate are controlled through the front panel of Alltech 2000ES, not Empower?
By B on Wednesday, August 18, 2004 - 07:18 am:
No, Empower will still control the pump, injector, etc. and it will also record the signal from the analog/digital convertor (SAT/IN unit for Waters). You will just have a different selection for the detector in your instrument method. All the panel on the detector will control are the detector parameters.
Your Waters rep will happily provide you with all the info you need to get their next boat payment, I mean provide you with the equipment that you need :).
By cytw98 on Wednesday, August 18, 2004 - 10:07 am:
A & B,
Thanks for the helpful info!