The facts that convinced us:
The autosampler shows reproducibility (typically we find 0.1% RSD @ 1ul at cycle times of 1min). The sales guy showed me an application where they did UV runs at 25 seconds cycle time. But peaks become too narrow for satisfying MRMs then.
Best thing is: the system is upgradable. 30kE for the frontend, runs at pressures up to 300 bar or so. If we decide, that we need 650 bar, we can upgrade the system for less than 10kE. And it plugs directly into Analyst. No hassle with drivers developed by the lc vendor.
The Drawbacks at other systems, that I know of:
Waters injector has a lot more carryover, and less precise injection. Limitation of flow. At the moment we can go up to 2x10ml /min. Should we install the higher pressure option, this decreases to 2x5.
Thermo: correct me if I am wrong: their Pump seems to be a low pressure side mixing system. Kicked out of our considerations immediately.
Agilent: Only one dampener. If you want to go for low gradient delay you have to use an undampened high pressure gradient flow.