Hi Ike
At first You have to create an appropriate processing method:
1, integrate the std injection, set the integration parameters
2, select the component tab in the proc. method
3, select OPTION/FILL TABLE FROM RESULT from the toolbar. The Empower copy the result table to the component table.
4,delete the integrated unknowns
5, type the name of yuor standards. In the fit colunm you can set the cal curve type, the internal std colunm you can set the internal std of each components, the cal curve reference col.determine the other compound that use for quantitation, you can use the relative response factor if it measured. Check the other colunms in this tab for other calculation.
Next step input the std mixture content input.
Select the injection in the project window, Alter sample
The alter sample find the amount button. click it.
find the copy from process method button. The Empower copy the proc. mthod component table. Type the amount of your component, save it. Integrate the std-s.