What are the general rules to pick an internal standard for LC-MS/MS? Are there any references available regarding this? Thank You.
-
- Posts: 232
- Joined: Tue Sep 07, 2004 7:03 pm
I'm afraid I don't have any references handy, but I can tell you what we do. The best choice is an isotopically labeled version of the analyte. With deuterium labels, be sure they are in positions on the molecule where they won't exchange for H in solution. With carbon-13 labels, you won't have to worry about this. Also, make sure there are enough labels so that your unlabeled analyte won't interfere due to its naturally occurring isotopes. The minimum number of labels you need will therefore depend on what elements are present in your molecule.
Labeled compounds are expensive, and sometimes the project manager won't want to pay for it. In this case, I use the following general rules for the ISTD:
(1) Has similar functionality, so that it ionizes in the same mode and hopefully the same functional groups are responsible for the ionization.
(2) Elutes close to your analyte, or ideally it coelutes*.
(3) If it coelutes, it must have an MRM transition unique from the analyte.
*(4) Alternately, if you use an isomer of your analyte as ISTD that has the same MRM transitions, then it must not coelute with your analyte.
(5) If you are adding ISTD to your samples prior to preparation (some people call this a "surrogate"), then it should have similar extraction recoveries.
These rules are not set in stone and depend on your purpose in using the internal standard, but the more of these you satisfy, the better internal standard you will have.
Labeled compounds are expensive, and sometimes the project manager won't want to pay for it. In this case, I use the following general rules for the ISTD:
(1) Has similar functionality, so that it ionizes in the same mode and hopefully the same functional groups are responsible for the ionization.
(2) Elutes close to your analyte, or ideally it coelutes*.
(3) If it coelutes, it must have an MRM transition unique from the analyte.
*(4) Alternately, if you use an isomer of your analyte as ISTD that has the same MRM transitions, then it must not coelute with your analyte.
(5) If you are adding ISTD to your samples prior to preparation (some people call this a "surrogate"), then it should have similar extraction recoveries.
These rules are not set in stone and depend on your purpose in using the internal standard, but the more of these you satisfy, the better internal standard you will have.
2 posts
Page 1 of 1
Who is online
In total there are 370 users online :: 0 registered, 0 hidden and 370 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 370 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 370 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science