Jasmine, one way of relating (linear) gradient and isocratic separations is to use the "average" k' of peaks during the gradient (this is the k' that peaks have at the midpoint of the column); that average is symbolized as k*.
Steeper gradients are equivalent to smaller k* values, which makes intuitive sense (shorter times, sharper peaks, poorer resolution).
If you want to get deeper into it, check the thread on
HPLC Methods development for Basic Compounds:
http://www.sepsci.com/chromforum/viewto ... highlight=
Uwe and I have gone around on this topic a couple of times, but if you look at the math, you see that his estimate that the gradient volume should be at least 10x the column volume for a wide range gradient is equivalent to saying the k* value should be at least 2. A k' value of at least 2 is, not coincidentally, the FDA recommended guidance for a validated HPLC method.