Error message--please help on Agilent 1100 with Empower

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Greetings,
I got this error message and got instrument failure. Did any one see this before? And why did this happen?
G1365B: Data invalid. At least for some diodes of the array, the data produced are invalid due to too much light at a diode or a gain setting.

When I got this error message, it happened that I changed my mobile phase from 0.05% perchloric acid/ACN to 0.1% formic acid/ACN. Is this just coincidence or formic acid does give me problem?

My chromatograms showed that I had elevated baseline for 0.1% formic acid/ACN comparing with 0.05% perchloric acid/ACN.

Condition:
Agilent 1100 drived by Empower
I injected 50:50 ACN/H2O 80 mcl.

Thanks a lot!

Did you re-balance (i.e., "zero") the detector after changing mobile phases? The absorbance characteristics of the two mobile phases are completely different.

Check the diagnostics on the detector - spectral intensity, dark current, etc.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.

I did some more investigation on this problem.

When I change to H2O/ACN mobile phase, that problem was gone. I did not re-balance anything between each mobile phase change.

The problem did not result in long time instrument failure. I can still run samples after restart everything.

The problem happens only with formic acid 0.1%/ACN at 1.3 min, which is about the void volume.

I think people are using formic acid quite often.

Do I need to set the peak width to be smaller (I had 0.1 min, 0.05 min)
Should I turn on the reference? (it was off when I had the problem)

best

You can solve this by changing the parameters "Margin for negative absorbance" to a higher value.
You may observed high negative absorbance at the beginning of the run.
An other crazy Empower message...
Attila

^ that, or it's just a reference issue - you need to gather your reference spectra from another part of the chromatogram (a quieter part).
Thanks,
DR
Image
5 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry