Conditions:
waters HPLC system with autosampler and online degaser, but no oven of the column that was exposed to the air.
Column: phenomenex Luna
Mobile Phase: Acetonitrile and Phosphrat buffer (pH=6.98, phosphric acid to adjust) gradient elution. Actonitrile from 27% to 37% at 0-15min, from 37% to 80% at 15-30min.
I do the same method with the same bottle of mobile phase. But at the morning and noon the peak change, that cause the bad peak resolution of peak 2 and 3. pleas see the Image:
I want to know what cause the peak shift.
I think the resone :
1. the temperature is changing. but I test it, the chang range in our lab is +/- 3 centigrade. It is a problem?
2. the pump is not good. I suspect the the ratio is not stable. ( the pump vaile has been changed)
Could you give me the other advise!