Willy » Tue Jun 24, 2008 2:15 am
Hi there,
By the term you saw only one peak with the new column do you mean that even by heating or extending the time of the analysis you didnt see the second peak on the new column? I have 2 answers:
1-Ive experienced one time a problem with old-new columns. We did validation of a method with 2 columns. One was relatively old and the other was new. Depending on the solvent analysed some of them tends to elute faster on old columns and later on new one. ( ill will not explain but they are few reasons this could happen with old column). So we needed to extend our analysis time to be able to see the second peak from the newer column.
2-From what you say ( 0.4 minutes in retention time difference) the possible problem would be that they might co-elute. If one of them is giving a high respond and the other gives a small peak that could be the problem. Check if the peak is well shape in the new column. One of my post was about that, i couldnt separate Heptane from TEA on one column and was well separated on another column from the same type-length.....
Im betting on my second answer!
Please give feedback
Hope that will help!
Willy the GC