Linearity Validation - Best Practice?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
When validating a HPLC method for linearity what is the usual best practice? Is it to simply inject the lowest conc. followed by the next highest conc., followed by the next highest conc. etc. or is it better (as I have done in the past) to randomise the order of injection of your linearity solutions?
-
- Posts: 136
- Joined: Mon Aug 30, 2004 6:41 am
Sometimes, simple can be the best.....
We inject QC samples at LOQ, medium and highest concentration level and calculate the difference between nominal and calculated concentrations. These shoulld fall within acceptable limits, pending your application. Furthermore, evaluate the difference between nominal and back-calculated concentrations of the calibration samples.
regards
regards
-
- Posts: 55
- Joined: Tue Aug 31, 2004 8:40 am
I've done duplicates by first going from low to high and then back from high to low. This challenges the method for etc. carryover.
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
I usually start at ~150% of what I'd expect for assay samples & go down through LOQ & LOD.
Thanks,
DR

DR

-
- Posts: 18
- Joined: Fri Oct 08, 2004 1:52 pm
When you have that wide of a range (LOQ <0.1%? - 150%) do you still have an even distribution of points? We have recently been discussing the validity of the typical least squares linear regression when estimating the line for a wide range with uneven distribution of points.
Any thoughts?
Thanks,
Any thoughts?
Thanks,
Ben
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I make a standard solution then take different aliquots to volume, to cover the expected range in the samples plus leeway. We take at least five dilutions (and a blank). At times we've gone high enough concentrations to document deviation from linearity.
6 posts
Page 1 of 1
Who is online
In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science